Maintenance Schedule & Procedures Teledyne API T703/T703U Calibrator Operation Manual
158
12. Once a lamp position is found that corresponds to a maximum observed value for
PHOTO_DET, tighten the lamp setscrew at the approximate maximum value
observed.
If the value of PHOTO_DET is not within the range of 4400 – 4600 mV, adjust it
accordingly.
13. Replace the cover on the instrument.